Maximum of 24/10/2012

Post Reply
Paolo Berardi
Posts: 578
Joined: Thu Sep 29, 2011 10:51 pm

Maximum of 24/10/2012

Post by Paolo Berardi »

Hi all, here is my last RR Lyr observation. Weather was very bad in the last weeks...

Relevant observation data:
- JD 2456225
- Date: 2012-10-24
- Observer: Paolo Berardi
- Exposure time: 300s
- Number of spectra: 31
- Duration: 2.5 hours
- Instrument: C9.25 - Lhires III 1200 - SXVR-H694
- WL range: 6330-6800 A
- R: ~5000

Animation series in cropped range 6500-6700 A:
Image


2d dynamic spectrum of h-alpha line in cropped range 6542-6581 A:
Image


Time evolution of h-alpha line (24 spectra in cropped range 6520-6605 A)
Image


Umberto Sollecchia simultaneous observation with Star Analyser 100:
Image


Ciao!
Paolo
Christian Buil
Posts: 1432
Joined: Mon Sep 26, 2011 6:59 pm
Contact:

Re: Maximum of 24/10/2012

Post by Christian Buil »

Paolo your spectra are really exceptionnal of quality!

Fine setup Lhires III + C9. Could you precise the slit width (23 microns, 35 microns, ...) ?

Christian B
Paolo Berardi
Posts: 578
Joined: Thu Sep 29, 2011 10:51 pm

Re: Maximum of 24/10/2012

Post by Paolo Berardi »

I'm glad for this Christian, your advices in OHP were very useful!
I'm quite happy with the setup. C9 is still easily manageable and it seems to me has a sufficent light collection.

For the efficiency I think most comes from the use of the sensor ICX694 in binning 2x2 mode and 35 micron slit (used up to now for my RR lyr observation). About resolution, I have noticed a considerable difference between the results obtained with the 35 um and 23 um slit aperture. But in my test in the same evening on AZ Cas I have detected a light loss of about 30%. I'm actually trying to capture an AZ Cas spectrum with 23 microns slit (2400 l/mm grating). When I reduce data I'll show you the result. I would like to know if, with AZ Cas, it's worth using the 23 um slit with my setup.

See you!
Paolo
Post Reply